Dear Grace:
I concur with Dr First. I have seen "reverse flow" many times . . . And in EVERY case, either the exhaust/roof blower was under-sized or the exhaust/roof blower was simply not operational! There are generally two alarms on BSC Class II B2s . . . One for the supply and one for the exhaust. Not sure which Baker model you have or which alarm(s) sounded!
Regards,
Paul J
Link to complete discussion "issues on defective biosafety cabinet": http://www.ghdonline.org/ic/discussion/issues-on-defective-biosafety-cabinet/
Grace Egos, MSPH
Hi Dr. Paul J.,
expand commentJust to give more info on the BSC and the blower and for everyone's comment:
BSC
Model: SG603TXCE
Serial Number :84326
Voltage : 230
Total Amperes : 11.3
Exhaust Blower
Name: Portable Ventilation
Type: JPV -300; Made in Taiwan
Static : 40/50 mm/aq
Power: 550watts
Speed : 2,900/3,320 rpm
Capacity :60/70 mm c.m.m
HZ : 50/60
Voltage:220 to 230v
The distance of the blower from the BSC would be about 2meters; there is a
low ceiling in that building so that the size of the duct is about 150mm x
600mm. I didn't have the chance to see the bend but you can see from the
pictures that there will be about 2 bends: first the duct that connects to
the BSC and the 2nd would be where the duct connects to the blower.
If you will look closely at the picture the duct is narrow compared to the
opening to the blower, (am not sure if I described it correctly).
Again thanks so much for the help and for connecting me to Baker.
Best regards,
Grace
-----Original Message-----
From: GHDonline (Paul A. Jensen, PhD, PE, CIH) [mailto ...
11:56 PM, 1 Oct 2008 | Permalink
Paul A. Jensen, PhD, PE, CIH
Dear Grace:
expand commentI spoke with Cindy from The Baker Company (www.bakerco.com) and I also found the specifications for the JPV-300 blower on the web (http://www.allproducts.com/machine/jouning/61-jpv200ss.html) . . . The maximum flow rate of the JPV-300 is 60 m3/min (2,100 cfm) at 40 mm w.g. (1.6 in w.g.). The BSC requires 15.4-16.0 m3/min (543-566 cfm) at 51-64 mm w.g. (2.0-2.5 in w.g.). The Baker’s round exhaust duct is 25.4 cm (10 in) in diameter. So, the area of the round duct is 508 cm2 (78.5 in2) and the area of the rectangular duct is 900 cm2 (139 in2).
So, here are my thoughts:
1. The initial duct size is OK, but not ideal.
2. The JPV-300 is a weak fan/blower and cannot handle the pressure loss of a well-designed system. With the duct transitions and elbow, there will be even greater losses. Hence, this fan does not pull enough air. The way to check this is to drill a hole in the duct, stick your velocity meter into the center of the duct, calculate the flow rate. I suspect ...
4:22 PM, 2 Oct 2008 | Permalink